Infrastructure tools to support an effective radiation oncology learning health system

Insulin pump
Paradigm insulin pump 530g, showing an infusion set loaded into spring-loaded insertion device called serter. A reservoir is attached to the infusion set (shown here removed from the pump).
Insulin pump in use
Diabetic child wearing a state-of-the-art insulin pump (referred to as a "patch pump"). His waterproof device needs no one infusion set

An insulin pump is a medical device used for the administration of insulin in the treatment of diabetes mellitus, also known as continuous subcutaneous insulin therapy. The device configuration may vary depending on design. A traditional pump includes:

  • the pump (including controls, processing module, and batteries)
  • a disposable reservoir for insulin (inside the pump)
  • a disposable infusion set, including a cannula for subcutaneous insertion (under the skin) and a tubing system to connect the insulin reservoir to the cannula.

Other configurations are possible. More recent models may include disposable or semi-disposable designs for the pumping mechanism and may eliminate tubing from the infusion set.

An insulin pump is an alternative to multiple daily injections of insulin by insulin syringes or an insulin pen and allows for flexible insulin therapy when used in conjunction with blood glucose monitoring and carbohydrate counting.

Medical uses

Insulin pumps are used to deliver insulin on a continuous basis to a person with type I diabetes.

Advantages

  • Users report better quality of life (QOL) compared to using other devices for administering insulin. The improvement in QOL is reported in type 1 and insulin-requiring type 2 diabetes subjects on pumps.[1]
  • The use of rapid-acting insulin for basal needs offers relative freedom from a structured meal and exercise regime previously needed to control blood sugar with slow-acting insulin.[2]
  • Programmable basal rates allow for scheduled insulin deliveries of varying amounts at different times of the day. This is especially useful in controlling events such as the dawn phenomenon resulting in fewer and less severe low blood sugar events during the night.[3]
  • Many users feel that bolusing insulin from a pump is more convenient and discreet than injection.[3][4]
  • Insulin pumps make it possible to deliver more precise amounts of insulin than can be injected using a syringe. This supports tighter control over blood sugar and hemoglobin A1c levels, reducing the chance of long-term complications associated with diabetes. This is predicted to result in a long-term cost savings relative to multiple daily injections.[5]
  • Many modern insulin pumps have a "bolus wizard" that calculates how much bolus insulin is needed, taking into account expected carbohydrate intake, blood sugar level, and still-active insulin.[6]
  • Insulin pumps can provide an accurate record of insulin usage through their history menus. On many insulin pumps, this history can be uploaded to a computer and graphed for trend analysis.[7]
  • Neuropathy is a troublesome complication of diabetes resistant to usual treatment. There are reports of alleviation or even total disappearance of resistant neuropathic pain with the use of insulin pumps.[8]
  • Recent studies of use of insulin pumps in Type 2 diabetes have shown profound improvements in HbA1c, sexual performance, and neuropathy pain.[9]

Disadvantages

Insulin pumps, cartridges, and infusion sets may be far more expensive than syringes used for insulin injection with several insulin pumps costing more than $6,000; necessary supplies can cost over $300.[3] Another disadvantage of insulin pump use is a higher risk of developing diabetic ketoacidosis if the pump malfunctions.[3] This can happen if the pump battery is discharged, if the insulin is inactivated by heat exposure, if the insulin reservoir runs empty, the tubing becomes loose and insulin leaks rather than being injected, or if the cannula becomes bent or kinked in the body, preventing delivery.[3] Therefore, pump users typically monitor their blood sugars more frequently to evaluate the effectiveness of insulin delivery.

  • Since the insulin pump needs to be worn most of the time, pump users need strategies to participate in activities that may damage the pump, such as rough sports and activities in the water. Some users may find that wearing the pump all the time (together with the infusion set tubing) is uncomfortable or unwieldy.
  • Possibility of insulin pump malfunctioning, and having to resort back to multiple daily injections until a replacement becomes available. However most pump manufacturers will have a program that will get a new pump to the user within 24 hours or allow the user to buy a second pump as a backup for a small fee. Additionally the pump itself will make many safety checks throughout the day, in some cases up to 4,000,000 and may have a second microprocessor dedicated to this.[citation needed]
  • Users may experience scar tissue buildup around the inserted cannula, resulting in a hard bump under the skin after the cannula is removed. The scar tissue does not heal particularly fast, so years of wearing the pump and changing the infusion site will cause the user to start running out of viable "spots" to wear the pump. In addition, the areas with scar tissue buildup generally have lower insulin sensitivity and may affect basal rates and bolus amounts. In some extreme cases the insulin delivery will appear to have no/little effect on lowering blood glucose levels and the site must be changed.
  • Users may experience allergic reactions and other skin irritation from the adhesive on the back of an infusion set. Experience may vary according to the individual, the pump manufacturer, and the type of infusion set used.
  • A larger supply of insulin may be required in order to use the pump. Many units of insulin can be wasted while refilling the pump's reservoir, filling the tubing, or changing an infusion site. This may affect prescription and dosage information.

Accessibility

Use of insulin pumps is increasing because of:

  • Easy delivery of multiple insulin injections for those using intensive insulin therapy.
  • Accurate delivery of very small boluses, helpful for infants.
  • Growing support among doctors and insurance companies due to the benefits contributing to reducing the incidence of long-term complications.
  • Improvements in blood glucose monitoring. New meters require smaller drops of blood, and the corresponding lancet poke in the fingers is smaller and less painful. These meters also support alternate site testing for the most routine tests for practically painless testing.

History

In the early 1960s, Dr. Arnold Kadish, a physician working at Loma Linda University, developed a closed-loop device for blood glucose control using a double-lumen autoanalyzer to measure blood glucose linked to an intravenous insulin infusion pump. Although it worked, the prototype was considered so bulky, worn as a backpack, and complicated, that it was never commercialized.[10] In 1976, the first commercialized insulin pump was created and was named the Biostator. The pump was a 60 kg bedside device.[11] It also had the capability of monitoring blood glucose levels, so this also doubles as the first continuous glucose monitor. Between 1978 and 1988, Robert Channon, working with Guy's Hospital and the Bristol Royal Infirmary, developed a series of miniature insulin infusion pumps.[12][13] Today, insulin pumps are so small that they can fit in a pocket or a purse.

In 1984, an Infusaid implantable infusion device was used to treat a 22-year-old patient successfully.[14]

The insulin pump was first endorsed in the United Kingdom in 2003 by the National Institute for Health and Care Excellence (NICE).

Developments

New insulin pumps are becoming "smart" as new features are added to their design. These simplify the tasks involved in delivering an insulin bolus.

  • insulin on board: This calculation is based on the size of a bolus, the time elapsed since the completion of the bolus, and a programmable metabolic rate. The pump software will estimate the insulin remaining in the bloodstream and relay it to the user. This supports the process of performing a new bolus before the effects of the last bolus are complete and, thereby, helps prevent the user from overcompensating for high blood sugar with unnecessary correction boluses.
  • bolus calculators: Pump software helps by calculating the dose for the next insulin bolus. The user enters the grams of carbohydrates to be consumed, and the bolus "wizard" calculates the units of insulin needed. It adjusts for the most recent blood glucose level and the insulin on board, and then suggests the best insulin dose to the user to approve and deliver.
  • custom alarms: The pump can monitor for activities during specific times of day and alarm the user if an expected activity did not occur. Examples include a missed meal bolus, a missed blood glucose test, a new blood glucose test 15 minutes after a low blood glucose test, etc. The alarms are customized for each user.
  • touch bolus: For persons with visual impairments, this button on the pump can be used to bolus for insulin without using the display. This works with a system of beeps to confirm the bolus parameters to the pump user. This feature is described as 'touch', 'audio', or 'easy' bolus depending on brand. The feature was first introduced in the mid- to late 1990s.[citation needed]
  • interface to personal computers: Since the late 1990s, most pumps can interface with personal computers for managing and documenting pump programming and/or to upload data from the pump. This simplifies record keeping and can be interfaced with diabetes management software.
  • integration with blood glucose meters: Blood glucose data can be manually entered into the pump to support the bolus wizard for calculation of the next insulin bolus. Some pumps support an interface between the insulin pump and a blood glucose meter.
    • The Medtronic Diabetes Minimed Paradigm series of insulin pumps allow for radio frequency (RF) communication. This enables the pump to receive data from a Lifescan (in the US) or Bayer (in other countries) blood glucose meter.
    • The Animas Ping is a pump/blood glucose meter combo that connect to each other using radio frequency. They both can work independently of each other and each have their own history storage. The main purpose of the connection between the pump and the meter is that it allows boluses to be made from the meter or the pump. This is particularly useful when correcting for a high blood sugar as the meter remembers readings and automatically enters them in correction boluses if they are less than 15 minutes old.
    • The DANA Diabecare IISG insulin pump has a blood glucose meter in it. After a blood glucose check with the integrated glucometer, the user can use the bolus wizard to deliver a required bolus.
    • The Insulet OmniPod has a separate remote, also known as a Personal Diabetes Monitor (PDM), that features a built-in meter that uses Freestyle test strips. This eliminates the need to carry and manage a separate meter or transfer blood glucose results from device to device.
  • integration with continuous glucose monitoring systems: Some insulin pumps can be used as a display for interstitial glucose values obtained from a continuous glucose monitoring system or sensor.
    • The Minimed Paradigm series' RF link also supports a continuous blood glucose sensor known as the Paradigm REAL-Time Continuous Glucose Monitor that wirelessly provides an interstitial glucose value every 5 minutes on the pump screen. The Medtronic REAL-Time System was the first to link a continuous monitor with an insulin pump system. In the Minimed 530G with Enlite (in the US) or Paradigm Veo (in other countries), the pump can enter a low glucose suspend mode stopping all insulin delivery (bolus and basal insulin) if interstitial glucose values fall below the hypoglycemia threshold. In the Minimed 640G insulin pump series, low glucose suspend mode can also be entered based on predicted hypoglycemia.[15]
    • The Animas Vibe is an insulin pump that is fully integrated with the Dexcom G4 Continuous Glucose Monitor. The two connect wirelessly to monitor and track blood glucose levels and detect patterns. The Dexcom G4 has the advantage of being designed to monitor glucose levels every five minutes throughout 7 days of continuous wear.[16] The Animas Vibe was approved for use in Europe in 2011, and Canada and the United States in January and December 2014, respectively.[17] NOTE: Animas insulin pumps are not available due to Johnson & Johnson's decision to cease operations at their Animas subsidiary.[18]
    • The Tandem Diabetes Care t:Slim X2 was approved by the U.S. Food and Drug Administration in 2019 and is the first insulin pump to be designated as an alternate controller enabled (ACE) insulin pump. ACE insulin pumps allow users to integrate continuous glucose monitors, automated insulin dosing (AID) systems, and other diabetes management devices with the pump to create a personalized diabetes therapy system.[19] Many users of the t:slim X2 integrate the pump with the Dexcom G6, a continuous glucose monitor approved by the FDA in 2018. It was the first CGM authorized for use in an integrated therapy system. The device does not require users to provide fingerstick calibrations and lasts for up to ten days.[20]
  • Other options may include remote control, tubeless pod, touch screen interface, rechargeable battery, pre-filled insulin cartridge, and mobile bolus.

MiniMed 670G is a type of insulin pump and sensor system created by Medtronic. It was approved by the US FDA in September 2016 and was the first approved hybrid closed loop system which senses a patient's basal insulin requirement and automatically adjusts its delivery to the body.[21][22][23]

Mylife YpsoPump, developed by Ypsomed, was launched in Europe in 2016.[24] Eli Lilly had planned to work with Ypsomed to introduce this to the United States, but that effort was terminated in 2022.[24]

Omnipod 5: On January 28, 2022, Insulet Corporation announced the FDA has approved the Omnipod 5, the first tubeless closed loop insulin pump with Smartphone control, working with the Dexcom G6 Continuous Glucose Monitor. The Omnipod 5 will have a feature named SmartAdjust technology that allows for the increase, decrease, or suspension of insulin based on the user's custom blood glucose targets.[25]

INSUL by AgVa: AgVa Healthcare announced that Insul by Agva is the world's most advanced and affordable Insulin pump. Features such as a built-in glucometer, Bluetooth connectivity, android and IOS app, as well as long-lasting and economical disposables.[26]

Future developments

  • When insulin pump technology is combined with a continuous blood glucose monitoring system, the technology seems promising for real time control of the blood sugar level. Currently[when?] there are no mature algorithms to automatically control the insulin delivery based on feedback of the blood glucose level. When the loop is closed, the system may function as an artificial pancreas.
  • Insulin pumps are being used for infusing pramlintide (brand name Symlin, or synthetic amylin) with insulin for improved postprandial glycemic control compared to insulin alone.
  • Dual hormone insulin pumps that infuse either insulin or glucagon. In event of hypoglycemia, the glucagon could be triggered to increase the blood glucose. This would be particularly valuable in a closed loop system under the control of a glucose sensor. The Artificial Pancreas, currently in clinical trials for FDA approval, is a recently developed device designed with this technology in mind.[27]
  • Ultrafast insulins. These insulins are absorbed more quickly than the currently available Humalog, Novolog, and Apidra which have a peak at about 60 minutes.[28] Faster insulin uptake would theoretically coordinate with meals better, and allow faster recovery from hyperglycemia if the insulin infusion is suspended. Biodel is developing an ultrafast insulin[29]

Dosing

Filling an insulin pump reservoir. (Left to right) 1: Reservoir in sterile packaging. 2: Filling the reservoir. 3: Reservoir with needle and plunger removed, ready for attachment to infusion set.

An insulin pump allows the replacement of slow-acting insulin for basal needs with a continuous infusion of rapid-acting insulin.

The insulin pump delivers a single type of rapid-acting insulin in two ways:[30]

  • a bolus dose that is pumped to cover food eaten or to correct a high blood glucose level.
  • a basal dose that is pumped continuously at an adjustable basal rate to deliver insulin needed between meals and at night.

Bolus shape

An insulin pump user can influence the profile of the rapid-acting insulin by shaping the bolus. Users can experiment with bolus shapes to determine what is best for any given food, which means that they can improve control of blood sugar by adapting the bolus shape to their needs.

A standard bolus is an infusion of insulin pumped completely at the onset of the bolus. It's the most similar to an injection. By pumping with a "spike" shape, the expected action is the fastest possible bolus for that type of insulin. The standard bolus is most appropriate when eating high carb low protein low fat meals because it will return blood sugar to normal levels quickly.

An extended bolus is a slow infusion of insulin spread out over time. By pumping with a "square wave" shape, the bolus avoids a high initial dose of insulin that may enter the blood and cause low blood sugar before digestion can facilitate sugar entering the blood. The extended bolus also extends the action of insulin well beyond that of the insulin alone. The extended bolus is appropriate when covering high fat high protein meals such as steak, which will be raising blood sugar for many hours past the onset of the bolus. The extended bolus is also useful for those with slow digestion (such as with gastroparesis or coeliac disease).

A combination bolus/multiwave bolus is the combination of a standard bolus spike with an extended bolus square wave. This shape provides a large dose of insulin up front, and then also extends the tail of the insulin action. The combination bolus is appropriate for high carb high fat meals such as pizza, pasta with heavy cream sauce, and chocolate cake.

A super bolus is a method of increasing the spike of the standard bolus. Since the action of the bolus insulin in the blood stream will extend for several hours, the basal insulin could be stopped or reduced during this time. This facilitates the "borrowing" of the basal insulin and including it into the bolus spike to deliver the same total insulin with faster action than can be achieved with spike and basal rate together. The super bolus is useful for certain foods (like sugary breakfast cereals) which cause a large post-prandial peak of blood sugar. It attacks the blood sugar peak with the fastest delivery of insulin that can be practically achieved by pumping.

Bolus timing

Since the pump user is responsible to manually start a bolus, this provides an opportunity for the user to pre-bolus to improve upon the insulin pump's capability to prevent post-prandial hyperglycemia. A pre-bolus is simply a bolus of insulin given before it is actually needed to cover carbohydrates eaten.

There are two situations where a pre-bolus is helpful:

  1. A pre-bolus of insulin will mitigate a spike in blood sugar that results from eating high glycemic foods. Infused insulin analogs such as NovoLog and Apidra typically begin to reduce blood sugar levels 15 or 20 minutes after infusion. As a result, easily digested sugars often hit the bloodstream much faster than infused insulin intended to cover them, and the blood sugar level spikes upward as a result. If the bolus were infused 20 minutes before eating, then the pre-bloused insulin would hit the bloodstream simultaneously with the digested sugars to control the magnitude of the spike.
  2. A pre-bolus of insulin can combine a meal bolus and a correction bolus when the blood sugar is above the target range before a meal. The timing of the bolus is a controllable variable to bring down the blood sugar level before eating again causes it to increase.

Similarly, a low blood sugar level or a low glycemic food might be best treated with a bolus after a meal is begun. The blood sugar level, the type of food eaten, and a person's individual response to food and insulin affect the ideal time to bolus with the pump.

Basal rate patterns

The pattern for delivering basal insulin throughout the day can also be customized with a pattern to suit the pump user.

  • A reduction of basal at night to prevent low blood sugar in infants and toddlers.
  • An increase of basal at night to counteract high blood sugar levels due to growth hormone in teenagers.
  • A pre-dawn increase to prevent high blood sugar due to the dawn effect in adults and teens.
  • In a proactive plan before regularly scheduled exercise times such as morning gym for elementary school children or after-school basketball practice for high school children.

Basal rate determination

Basal insulin requirements will vary between individuals and periods of the day. The basal rate for a particular time period is determined by fasting while periodically evaluating the blood sugar level. Neither food nor bolus insulin must be taken for 4 hours before or during the evaluation period. If the blood sugar level changes dramatically during evaluation, then the basal rate can be adjusted to increase or decrease insulin delivery to keep the blood sugar level approximately steady.

For instance, to determine an individual's morning basal requirement, they must skip breakfast. On waking, they would test their blood glucose level periodically until lunch. Changes in blood glucose level are compensated with adjustments in the morning basal rate. The process is repeated over several days, varying the fasting period, until a 24-hour basal profile has been built up which keeps fasting blood sugar levels relatively steady. Once the basal rate is matched to the fasting basal insulin need, the pump user will then gain the flexibility to skip or postpone meals such as sleeping late on the weekends or working overtime on a weekday.

Many factors can change insulin requirements and require an adjustment to the basal rate:

  • continued beta cell death following diagnosis of type 1 diabetes (honeymoon period)
  • growth spurts particularly during puberty
  • weight gain or loss
  • any drug treatment that affects insulin sensitivity (e.g. corticosteroids)
  • eating, sleeping, or exercise routine changes
  • whenever the control over hyperglycemia is degrading
  • and according to the seasons.

A pump user should be educated by their diabetes care professional about basal rate determination before beginning pump therapy.

Temporary basal rates

Since the basal insulin is provided as a rapid-acting insulin, the basal insulin can be immediately increased or decreased as needed with a temporary basal rate. Examples when this is helpful include:

  • As a passenger during a long car drive, when insulin needs are different due to inactivity.
  • While driving on an extended trip, to reduce the risk of hypoglycemia, a lower temporary basal rate may be programmed.
  • During and after spontaneous exercise or sports activities, when the body needs less insulin.
  • During illness or stress, when basal demand increases due to insulin resistance.
  • When blood ketones are present, when additional insulin is needed.
  • When on an extended fast (such as Ramadan, Lent, or Yom Kippur) when basal requirements may be lower.
  • During menses, when additional basal insulin might be needed.

Security

In August 2011, an IBM researcher, Jay Radcliffe, demonstrated a security flaw in insulin pumps. Radcliffe was able to hack the wireless interface used to control the pump remotely.[31] Pump manufacturer Medtronic later said security research by McAfee uncovered a flaw in its pumps that could be exploited.[32]

See also

References

  1. ^ Kesavadev J, Kumar A, Ahammed S, Jothydev S (2008). "Experiences with Insulin Pump in 52 Patients with Type 2 Diabetes in India". DiabetesPro. American Diabetes Association. 2021-PO. Archived from the original on 24 February 2012.
  2. ^ Muppidi R. "Insulin Pump Training". Advanced Endocrine and Diabetes Hospital. AED Hospital. Retrieved 3 December 2019.
  3. ^ a b c d e Millstein R, Becerra NM, Shubrook JH (December 2015). "Insulin pumps: Beyond basal-bolus". Cleveland Clinic Journal of Medicine (Review). 82 (12): 835–842. doi:10.3949/ccjm.82a.14127. PMID 26651892.
  4. ^ Graveling AJ, McIntyre EA (January 2009). "Insulin Delivery Devices". Journal of the Royal College of Physicians of Edinburgh. 39 (2): 146–150.
  5. ^ Conget Donlo I, Serrano Contreras D, Rodríguez Barrios JM, Levy Mizrahi I, Castell Abat C, Roze S (2006). "[Cost-utility analysis of insulin pumps compared to multiple daily doses of insulin in patients with type 1 diabetes mellitus in Spain]". Revista Espanola de Salud Publica (in Spanish). 80 (6): 679–695. doi:10.1590/s1135-57272006000600008. PMID 17147307.
  6. ^ Zisser H, Wagner R, Pleus S, Haug C, Jendrike N, Parkin C, et al. (December 2010). "Clinical performance of three bolus calculators in subjects with type 1 diabetes mellitus: a head-to-head-to-head comparison". Diabetes Technology & Therapeutics. 12 (12): 955–961. doi:10.1089/dia.2010.0064. PMID 21128842.
  7. ^ Steineck I, Cederholm J, Eliasson B, Rawshani A, Eeg-Olofsson K, Svensson AM, et al. (June 2015). "Insulin pump therapy, multiple daily injections, and cardiovascular mortality in 18,168 people with type 1 diabetes: observational study". BMJ. 350 (jun22 1): h3234. doi:10.1136/bmj.h3234. PMC 4476263. PMID 26100640.
  8. ^ Kesavadev J, Rasheed SA. "Dramatic Response of Painful Peripheral Neuropathy with Insulin Pump in Type 2 Diabetes". DiabetesPro. American Diabetes Association. 2097-PO. Archived from the original on 24 February 2012.
  9. ^ Kesavadev J, Balakrishnan S, Ahammed S, Jothydev S (August 2009). "Reduction of glycosylated hemoglobin following 6 months of continuous subcutaneous insulin infusion in an Indian population with type 2 diabetes". Diabetes Technology & Therapeutics. 11 (8): 517–521. doi:10.1089/dia.2008.0128. PMID 19698065.
  10. ^ March C, Libman I, Becker D, Levitsky L (November 2022). "From Antiquity to Modern Times: A History of Diabetes Mellitus and Its Treatments" (PDF). Hormone Research in Paediatrics: 602. PMID 36446320. Retrieved 11 December 2024.
  11. ^ Mascini M (April 2007). A Brief Story of Biosensor Technology, chapter in Biotechnological Applications of Photosynthetic Proteins: Biochips, Biosensors and Biodevices. pp. 4–10. Retrieved 24 March 2023.
  12. ^ "Insulin pump calls the shots". New Scientist. Vol. 88, no. 1226. 6 November 1980. p. 369. Retrieved 24 March 2023.
  13. ^ "Multiple injection infusion device (portable), UK Patent Application GB2222525A, 28 March 1988" (PDF). Google Patents. Retrieved 24 March 2023.
  14. ^ Campbell IW, Kritz H, Najemnik C, Hagmueller G, Irsigler K (July 1984). "Treatment of type I diabetic with subcutaneous insulin resistance by a totally implantable insulin infusion device ("Infusaid")". Diabetes Research. 1 (2): 83–88. PMID 6442226.
  15. ^ "What is SmartGuard™ Technology?". 12 March 2018. Archived from the original on 21 June 2015. Retrieved 21 June 2015.
  16. ^ "Animas Vibe and CGM-system". Animas. Archived from the original on 23 January 2014. Retrieved 28 January 2014.
  17. ^ "Animas Vibe Insulin Pump with Latest Dexcom CGM Technology Now Available in Canada". CNW. Archived from the original on 16 July 2015. Retrieved 28 January 2014.
  18. ^ Brown A (17 October 2017). "Animas Closes Operations and Exits Insulin Pump Market". Diatribe.org. Retrieved 24 March 2018.
  19. ^ "FDA authorizes first interoperable insulin pump intended to allow patients to customize treatment through their individual diabetes management devices". FDA. 14 February 2019. Retrieved 4 March 2021.
  20. ^ "FDA authorizes first fully interoperable continuous glucose monitoring system, streamlines review pathway for similar devices". FDA. 27 March 2018. Retrieved 4 March 2021.
  21. ^ "Recently approved devices: The 670G System - P160017". FDA. 28 September 2016.
  22. ^ a. Dealing with any kind of medical technology, there are going to be pros and cons. With an insulin pump there are many to consider. Some of the pros of insulin pump therapy are precise insulin delivery down to the 0.025 minimum. They also replace the need to give a shot each time you eat. It allows for easier exercise management. Another large pro to an insulin pump is the reduced chance of variability with the patient’s blood glucose levels. Some of the cons are the cost of the insulin pump and supplies, the risk of infection at the pump site, and risk of DKA because of a pump malfunction.
  23. ^ "What are the Advantages and Disadvantages of an Insulin Pump?". Children's Hospital of Michigan. 18 November 2019. Retrieved 8 May 2021.
  24. ^ a b Whooley, Sean (9 December 2022). "Eli Lilly discontinues Ypsomed collaboration to pursue its own U.S. insulin pump offering". Medical Design & Outsourcing. Retrieved 10 December 2022.
  25. ^ "Omnipod® 5 | Omnipod". www.omnipod.com. Retrieved 28 January 2022.
  26. ^ "INSUL by AgVa | AgVa Healthcare". Retrieved 29 March 2022.
  27. ^ Brown A, Liu N (31 March 2014). "Dr. Ed Damiano Presents Next Set of Bionic Pancreas Study Results at ATTD". diaTribe. The diaTribe Foundation. Retrieved 19 March 2015.
  28. ^ "Humalog prescribing information" (PDF). Eli Lilly and Company. 2019.
  29. ^ "Linjeta duration of action". Biodel, Inc. Archived from the original on 31 March 2013.
  30. ^ "Insulin pumps". Diabetes.co.uk. 15 January 2019.
  31. ^ "Insulin Pumps Vulnerable to Hacking". Fox News. Associated Press. 22 October 2015. Archived from the original on 5 August 2011.
  32. ^ "Exclusive: Medtronic probes insulin pump risks". Reuters. 25 October 2011.